Troubleshooting and Service

9.4.11.3. Status Outputs

260
To test the status output electronics:
1. Connect a jumper between the "D" pin and the "" pin on the status output
connector.
2. Connect a 1000 ohm resistor between the "+" pin and the pin for the status output
that is being tested.
3. Connect a voltmeter between the "" pin and the pin of the output being tested (see
table below).
4. Under the DIAG Signal I/O menu (See Section9.1.3), scroll through the inputs and
outputs until you get to the output in question.
5. Alternatively, turn on and off the output noting the voltage on the voltmeter.
It should vary between 0 volts for ON and 5 volts for OFF.

Table 9-12: Status Outputs Check

PIN (LEFT TO RIGHT)
1
2
3
4
5
6
7
8
Teledyne API – Model T700 Dynamic Dilution Calibrator
STATUS
ST_SYSTEM_OK
SPARE
ST_CAL_ACTIVE
ST_DIAG_MODE
ST_TEMP_ALARM
ST_PRESS_ALARM
PERM_VALVE_1
PERM_VALVE_2
06873B DCN6388