Noisy, Unstable, Or Non-Linear Span O Readings; Slow Response To Changes In Concentration; The Analog Output Signal Level Does Not Agree With Front Panel Readings; Cannot Zero - Teledyne T700 Operation Manual

Dynamic dilution calibrator
Hide thumbs Also See for T700:
Table of Contents

Advertisement

Troubleshooting and Service
9.5.1.2. Noisy, Unstable, or Non-Linear Span O

9.5.1.3. Slow Response to Changes in Concentration

9.5.1.4. The Analog Output Signal Level Does Not Agree With Front Panel Readings

9.5.1.5. Cannot Zero

9.5.1.6. Cannot Span

266
Check for leaks in the pneumatic systems as described in Section 8.2.1.
Check for proper operation of the meas/ref switching valve as described in Section
9.5.2.
Check for dirty absorption cell and clean or replace as necessary as described in
Section 8.2.2.
Check for operation of the A/D circuitry on the motherboard. See Section 9.4.11.1.
Confirm the Sample Temperature, Sample Pressure and Sample Flow readings are
correct. Check and adjust as required.
Check for dirty absorption cell and clean or replace as necessary as described in
Section 8.2.2.
Check for pneumatic leaks as described in Section 8.2.1.
The photometer needs 800 cm
when calculating total required output flow for the calibrator (see Section 3.4.9).
Confirm that the recorder offset (see Section 4.10.1.5) is set to zero.
Perform an AIO calibration (see Section 4.10.1.6) and photometer dark calibration
(see Section 7.3.5).
Check for leaks in the pneumatic system as described in Section 8.2.1.
Confirm that the Zero gas is free of Ozone.
The photometer needs 800 cm
when calculating total required output flow for the calibrator (see Section 3.4.9).
Check for leaks in the pneumatic systems as described in Section 8.2.1.
Check for proper operation of the meas/ref switching valve as described in
Section9.5.2.
Check for dirty absorption cell and clean or replace as necessary as described in
Section 8.2.2.
Check for operation of the A/D circuitry on the motherboard. See Section 9.4.11.1.
Confirm the Sample Temperature, Sample Pressure and Sample Flow readings are
correct. Check and adjust as required.
The photometer needs 800 cm
when calculating total required output flow for the calibrator (see Section 3.4.9).
Teledyne API – Model T700 Dynamic Dilution Calibrator
Readings
3
3
/min of gas flow. Ensure that this is accounted for
3
/min of gas flow. Ensure that this is accounted for
3
/min of gas flow. Ensure that this is accounted for