Specifications And Approvals; Specifications; Table 2-1: Model T360 Basic Unit Specifications - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

SPECIFICATIONS AND APPROVALS

2.

2.1. SPECIFICATIONS

T360 Parameter
Ranges
(Physical Analog Output)
Measurement Units
Zero Noise
Span Noise
Lower Detectable Limit
Zero Drift (24 hours)
Span Drift (24 hours)
Lag Time
Rise/Fall Time
Linearity
Precision
Sample Flow Rate
Temperature Coefficient
Voltage Coefficient
AC Power Rating
Analog Output Ranges
Analog Output Resolution
Recorder Offset
Standard I/O
Optional I/O
Environmental
Temperature Range
Humidity Range
Dimensions H x W x D
Weight
1
At constant temperature and voltage.
07272B DCN6552

Table 2-1: Model T360 Basic Unit Specifications

T360 Specification
Min: 0-2 ppm Full scale
Max: 0-2,000 ppm Full scale
Selectable, dual ranges and auto ranging supported
3
ppb, ppm, µg/m
, mg/m
< 0.1 ppm (RMS)
< 1% of reading (RMS)
1
< 0.2 ppm
1
<0.25 ppm
1
<0.5% of reading
10 seconds
<60 seconds to 95%
1% of full scale
0.5% of reading
3
800cm
/min. ±10%
O
Sensor option adds 110 cm³/min, ±20%, to total flow through when installed.
2
< 0.1% of Full Scale per
< 0.05% of Full Scale per V
100V-120V, 60 Hz, (142 W) ; 220V – 240 V, 50 Hz (147 W)
10V, 5V, 1V, 0.1V (selectable)
1 part in 4096 of selected full-scale voltage
±10%
1 Ethernet: 10/100Base-T
2 RS-232 (300 – 115,200 baud)
2 USB device ports
8 opto-isolated digital status outputs
6 opto-isolated digital control inputs (3 defined, 3 spare)
4 analog outputs
1 USB com port
1 RS485
8 analog inputs (0-10V, 12-bit)
4 digital alarm outputs
Multidrop RS232
3 4-20mA current outputs
Installation category (over-voltage category) II; Pollution degree 2
o
5-40
C
0 - 95% RH, non-condensing
7" x 17" x 23.5" (178 mm x 432 mm x 597 mm)
40 lbs. (18.1 kg)
3
, % (user selectable)
o
C
21

Advertisement