Teledyne API Model T360/T360M Operation Manual
7.4.4.2. SIGNAL SYNCHRONIZATION AND DEMODULATION
DESIGNATION
Measure Gate
Measure Dark Gate
Reference Gate
Reference Dark Gate
07272B DCN6552
The signal emitted by the IR photo-detector goes through several stages of
amplification before it can be accurately demodulated.
amplification stage that raises the signal to levels readable by the rest of the
synch/demod board circuitry. The second is a variable amplification stage that is
adjusted at the factory to compensate for performance variations of mirrors,
detectors, and other components of the optical bench from instrument to
instrument.
The workhorses of the sync/demod board are the four sample-and-hold circuits
that capture various voltage levels found in the amplified detector signal needed
to determine the value of CO2 MEAS and CO2 REF. They are activated by
logic signals under the control of a compact programmable logic device (PLD),
which in turn responds to the output of the Segment Sensor and M/R Sensor
depicted in Figure 7-14.
The four sample and hold circuits are:
IR BEAM PASSING THROUGH
MEASUREMENT cell of GFC Wheel
MEASUREMENT cell of GFC Wheel
REFERENCE cell of GFC Wheel
REFERENCE cell of GFC Wheel
Timing for activating the Sample and Hold circuits is provided by a phase lock
loop circuit (PLL). Using the segment sensor output as a reference signal the
PLL generates clock signal at ten times that frequency. This faster clock signal is
used by the PLD to make the sample and hold circuits capture the signal during
the center portions of the detected waveform, ignore the rising and falling edges
of the detector signal.
Figure 7-16:
ACTIVE WHEN:
SEGMENT SENSOR PULSE IS:
Sample & Hold
Active
Sample & Hold
Inactive
Sample & Hold Timing
Principles of Operation
The first is a pre-
HIGH
LOW
HIGH
LOW
Detector
Output
211