Table 3-12: Possible Warning Messages At Start-Up - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Getting Started
MESSAGE
ANALOG CAL WARNING
BENCH TEMP WARNING
BOX TEMP WARNING
CANNOT DYN SPAN
CANNOT DYN ZERO
CONC ALRM1 WARNING
CONC ALRM2 WARNING
CONFIG INITIALIZED
DATA INITIALIZED
O2 CELL TEMP WARN
PHOTO TEMP WARNING
REAR BOARD NOT DET
RELAY BOARD WARN
SAMPLE FLOW WARN
SAMPLE PRESS WARN
SAMPLE TEMP WARN
SOURCE WARNING
SYSTEM RESET
WHEEL TEMP WARNING
56
Table 3-12 lists brief descriptions of the warning messages that may occur during
start up.

Table 3-12: Possible Warning Messages at Start-Up

The instrument's A/D circuitry or one of its analog outputs is not calibrated.
The optical bench Temperature is outside the specified limits.
Remote span calibration failed while the dynamic span feature was set to
turned on.
Remote zero calibration failed while the dynamic zero feature was set to
turned on.
Configuration was reset to factory defaults or was erased.
Concentration alarm 1 is enabled and the measured CO
point.
Concentration alarm 2 is enabled and the measured CO
point.
Configuration storage was reset to factory configuration or erased.
DAS data storage was erased.
O
sensor cell temperature outside of warning limits.
2
The temperature of the IR photometer is outside the specified limits.
The CPU is unable to communicate with the motherboard.
The firmware is unable to communicate with the relay board.
The flow rate of the sample gas is outside the specified limits.
Sample gas pressure outside of
The temperature of the sample gas is outside the specified limits.
The IR source may be faulty.
The computer was rebooted.
The Gas Filter Correlation wheel temperature is outside the specified limits.
Teledyne API Model T360/T360M Operation Manual
MEANING
operational parameters.
level is ≥ the set
2
level is ≥ the set
2
07272B DCN6552