Troubleshooting and Service
2
8.5.3. I
C Bus
8.5.4. Touchscreen Interface
248
A voltmeter should be used to verify that the DC voltages are correct per the
values in the table below, and an oscilloscope, in AC mode, with band limiting
turned on, can be used to evaluate if the supplies are producing excessive noise
(> 100 mV p-p).
Table 8-7: DC Power Supply Acceptable Levels
POWER
SUPPLY
VOLTAGE
FROM TEST POINT
ASSY
PS1
+5
PS1
+15
PS1
-15
PS1
Agnd
PS1
Chassis
PS2
+12
PS2
Dgnd
2
Operation of the I
C bus can be verified by observing the behavior of D1 on the
Relay PCA & D2 on the Valve Driver PCA . Assuming that the DC power
supplies are operating properly, the I
relay PCA and D2 of the Valve Driver PCA are flashing
There is a problem with the I
the Valve Driver PCA are ON/OFF constantly (steady lit or steady off)
Verify the functioning of the touchscreen by observing the display when pressing
a touchscreen control button. Assuming that there are no wiring problems and
that the DC power supplies are operating properly, if pressing a control button on
the display does not change the display, any of the following may be the problem:
The touchscreen controller may be malfunctioning.
The internal USB bus may be malfunctioning.
You can verify this failure by logging on to the instrument using APICOM or a
terminal program to any of the communications ports. If the analyzer responds to
remote commands and the display changes accordingly, the touchscreen interface
may be faulty.
Teledyne API Model T360/T360M Operation Manual
CHECK RELAY BOARD TEST POINTS
TO TEST POINT
NAME
#
NAME
Dgnd
1
Agnd
3
Agnd
3
-15V
Agnd
3
Dgnd
Dgnd
1
Chassis
+12V Ret
6
+12V
+12V Ret
6
Dgnd
2
C bus is operating properly if: D1 on the
2
C bus if both D1 on the relay PCA and D2 of
MIN V
#
+5
2
4.8
+15
4
13.5
5
-14V
1
-0.05
N/A
-0.05
7
11.75
1
-0.05
07272B DCN6552
MAX V
5.25
16V
-16V
0.05
0.05
12.5
0.05