Inability To Zero - Touchscreen Zero Button Not Visible - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
8.3.4. Inability to Zero – Touchscreen ZERO Button Not Visible
07272B DCN6552
1. Confirm that there is a good source of zero air. Dilute a tank of span gas
with the same amount of zero air from two different sources. If the CO
Concentration of the two measurements is different, there is a problem with
one of the sources of zero air.
2. Check for leaks in the pneumatic systems as described in Section 6.3.3.
3. If the analyzer has had zero/span valve options, the CO
maintenance.
4. Check to make sure that there is no ambient air leaking into zero air line.
Troubleshooting and Service
2
scrubber may need
2
243