Table 3-2: Rear Panel Description - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Getting Started
Component
cooling fan
AC power
connector
Model/specs label
PURGE IN
SAMPLE
EXHAUST
VENT SPAN
PRESSURE SPAN
RX TX
COM 2
RS-232
DCE DTE
STATUS
ANALOG OUT
CONTROL IN
ALARM
ETHERNET
ANALOG IN
Information Label
32

Table 3-2: Rear Panel Description

Pulls ambient air into chassis through side vents and exhausts through rear
Connector for three-prong cord to apply AC power to the analyzer
CAUTION! The cord's power specifications (specs) MUST comply with the power
specs on the analyzer's rear panel Model number label
Identifies the analyzer model number and provides power specs
Connect a source of dried air that has been scrubbed of CO
purge air to the GFC wheel housing (Section 7.1.3.1). (Also see Section 8.2.3).
Connect a gas line from the source of sample gas here.
Calibration gases are also inlet here on units without zero/span valve options installed.
Connect an exhaust gas line of not more than 10 meters long here that leads outside
the shelter or immediate area surrounding the instrument.
Span gas vent outlet for units with zero/span valve options installed.
Connect an exhaust gas line of not more than 10 meters long here.
On units with zero/span valve options installed, connect a gas line to the source of
calibrated span gas here.
Internal Zero Air: On units with zero/span valve options installed but no internal zero
IZS
air scrubber, attach a gas line to the source of zero air here.
LEDs indicate receive (RX) and transmit (TX) activity on the when blinking.
Serial communications port for RS-232 or RS-485.
Serial communications port for RS-232 only.
Switch to select either data terminal equipment or data communication equipment
during RS-232 communication.
For outputs to devices such as Programmable Logic Controllers (PLCs).
For voltage or current loop outputs to a strip chart recorder and/or a data logger.
For remotely activating the zero and span calibration modes.
Option for concentration alarms and system warnings.
Connector for network or Internet remote communication, using Ethernet cable.
Option for external voltage signals from other instrumentation and for logging these
signals.
Com port optional connector for direct connection to laptop computer, using USB cable.
USB
Includes voltage and frequency specifications.
Teledyne API Model T360/T360M Operation Manual