Range Units - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual

4.8.6. Range Units

Concentrations displayed in mg/m3 and ug/m3 use 0  C, 760 mmHg for Standard Temperature and
Pressure (STP). Consult your local regulations for the STP used by your agency.
Select the preferred
concentration unit.
Once the units of measurement have been changed the unit MUST be recalibrated, as the
"expected span values" previously in effect will no longer be valid. Simply entering new expected
span values without running the entire calibration routine is not sufficient.
07272B DCN6552
The T360 can display concentrations in parts per billion (10
6
parts per million (10
UG), milligrams per cubic meter (mg/m
sample gas, %). Changing units affects all of the display, analog outputs, COM
port and DAS values for all reporting ranges regardless of the analyzer's range
mode.
Conversion factors from volumetric to mass units are:
CO
: ppb x 1.96 = µg/m3; ppm x 1.96 = mg/m3
2
To change the concentration units:
SAMPLE
< TST TST > CAL
SETUP X.X
CFG DAS
RNGE
SETUP X.X
MODE SET
UNIT
SETUP X.X
PPM PPB
UGM
SETUP X.X
CONC UNITS: %
PPM PPB UGM MGM
mols per mol, PPM), micrograms per cubic meter (µg/m
3
, MG) or percent (volume CO
NOTE
RANGE = 500.00 PPB
CO2=X.XXX
PRIMARY SETUP MENU
PASS CLK MORE
RANGE CONTROL MENU
CONC UNITS: PPM
MGM
%
ENTER
%
ENTER EXIT
NOTE
Operating Instructions
9
mols per mol, PPB),
SETUP
EXIT
EXIT returns
to the main menu.
EXIT
EXIT
ENTR accepts
the new unit,
EXIT returns
to the SETUP
menu.
3
,
/volume
2
101