4.13.2. Signal I/O

The signal I/O diagnostic mode allows reviewing and changing the digital and
analog input/output functions of the analyzer. See Appendix A-4 for a complete
list of the parameters available for review under this menu.
Any changes of signal I/O settings will remain in effect only until the signal I/O menu is exited.
Exceptions are the ozone generator override and the flow sensor calibration, which remain as
To enter the signal I/O test mode, press:
SAMPLE
< TST TST > CAL
SETUP X.X
CFG DAS RNGE PASS CLK
EXIT returns
to the main
SAMPLE display
SETUP X.X
COMM VARS
SAMPLE
8
07272B DCN6552
entered when exiting.
RANGE = 500.000 PPM
CO2 =X.XXX
SETUP
PRIMARY SETUP MENU
MORE
EXIT
SECONDARY SETUP MENU
DIAG
ALRM
EXIT
ENTER SETUP PASS: 818
1
8
ENTR
EXIT
NOTE
DIAG
SIGNAL I / O
PREV NEXT
JUMP
DIAG I / O
Test Signals Displayed Here
PREV NEXT
JUMP
EXAMPLE
DIAG I / O
JUMP TO: 12
1
2
DIAG I / O
ST_CONC_VALID = ON
PREV NEXT JUMP
Pressing PRNT will send a formatted printout to the serial port and can be captured
Operating Instructions
Use NEXT & PREV to
move between signal
ENTR
EXIT
Use JUMP to go
directly to a specific
signal location
PRNT EXIT
See Appendix A-4 for
a complete list of
available SIGNALS
ENTR
EXIT
Enter 12 to Jump to
12) ST_CONC_VALID
ON
PRNT
EXIT
with a computer or other output device.
133
types.
EXAMPLE:
Exit to return
to the
DIAG menu