Oxygen Sensor (Opt 65); Paramagnetic Measurement Of O - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Principles of Operation
7.1.4.1. AMBIENT CO

7.2. OXYGEN SENSOR (OPT 65)

7.2.1. Paramagnetic Measurement of O

198
INTERFERENCE REJECTION
2
CO
absorbs IR light so well that even the narrow volume of ambient air between
2
the IR source and the sample chamber is enough to alter the analyzer's measured
concentration of CO
. Also, ambient air, which averages around 350 ppm to 400
2
ppm, will vary significantly over the course of the day. The ambient CO
concentration can rise as high as 1 000 ppm during the time of the day when
people are present. It can fluctuate  300 ppm as the photosynthesis of plant life
in the nearby area increases during the day and decreases at night.
The basic design of the T360 rejects most of this interference at a 100:1 ratio;
however this still can allow small fluctuations in CO
course of the day. To completely remove all effects of ambient CO
analyzer's measurement of CO
GFC wheel housing to purge all ambient CO
The oxygen sensor used in the T360 analyzer utilizes the fact that oxygen is
attracted into strong magnetic field; most other gases are not, to obtain fast,
accurate oxygen measurements.
The sensor's core is made up of two nitrogen filled glass spheres, which are
mounted on a rotating suspension within a magnetic field (Figure 7-7). A mirror
is mounted centrally on the suspension and light is shone onto the mirror that
reflects the light onto a pair of photocells. The signal generated by the photocells
is passed to a feedback loop, which outputs a current to a wire winding (in effect,
a small DC electric motor) mounted on the suspended mirror.
Oxygen from the sample stream is attracted into the magnetic field displacing the
nitrogen filled spheres and causing the suspended mirror to rotate. This changes
the amount of light reflected onto the photocells and therefore the output levels of
the photocells. The feedback loop increases the amount of current fed into the
winding in order to move the mirror back into its original position. The more O
present, the more the mirror moves and the more current is fed into the winding
by the feedback control loop.
A sensor measures the amount of current generated by the feedback control loop
which is directly proportional to the concentration of oxygen within the sample
gas mixture.
Teledyne API Model T360/T360M Operation Manual
, dried air, scrubbed of all CO
2
.
2
2
concentration during the
2
from the
2
is pumped into the
2
07272B DCN6552