Table 2-2: Model T360M Basic Unit Specifications - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Specifications and Approvals
T360M Parameter
Ranges
(Physical Analog Output)
Measurement Units
Zero Noise
Span Noise
Lower Detectable Limit
Zero Drift (24 hours)
Span Drift (24 hours)
Lag Time
Rise/Fall Time
Linearity
Precision
Sample Flow Rate
Temperature Coefficient
Voltage Coefficient
AC Power Requirements
Analog Output Ranges
Recorder Offset
Analog Output Resolution
Standard I/O
Optional I/O
Operating Temperature Range
Humidity Range
Dimensions H x W x D
Weight
Environmental
1
At constant temperature and voltage.
22

Table 2-2: Model T360M Basic Unit Specifications

T360M Specification
Min: 0-4 ppm Full scale
Max: 0-4000 ppm Full scale
Selectable, dual ranges and auto ranging supported
ppb, ppm, µg/m
< 0.2 ppm (RMS)
< 1% of reading (RMS)
1
< 0.4 ppm
1
<0.5 ppm
<0.5% of reading
10 seconds
<60 seconds to 95%
1% of full scale
0.5% of reading
3
800cm
/min. ±10%
O
Sensor option adds 110 cm³/min, ±20%, to total flow though when installed
2
< 0.1% of Full Scale per
< 0.05% of Full Scale per V
100V – 120V, 220V – 240V, 50/60 Hz
10V, 5V, 1V, 0.1V
±10%
1 part in 4096 of selected full-scale voltage
1 Ethernet: 10/100Base-T
2 RS232 (300-115,200 baud)
2 USB device ports
8 Status opto-isolated digital status outputs
6 Opto-isolated digital control inputs (3 defined, 3 spare)
4 Analog outputs
1 USB com port
1 RS485
8 Analog inputs (0-10V, 12-bit)
4 Digital alarm outputs
Multidrop RS232
3 4-20mA current outputs
o
5-40
C
0 - 95% RH, non-condensing
7" x 17" x 23.5" (178 mm x 432 mm x 597 mm)
40 lbs. (18.1 kg)
Installation category (over-voltage category) II; Pollution degree 2
Teledyne API Model T360/T360M Operation Manual
3
3
, mg/m
, (selectable)
1
o
C or 0.1 ppm per
o
C, whichever is greater
07272B DCN6552