Setting up Control Outputs for a Sequence
The calibrator's control outputs allow setting up the calibrator to control devices that accept
logic-level digital inputs, such as programmable logic controllers (PLCs), data loggers, or
digital relays/valve drivers.
They can be used as:
•
•
They can be set so that they are:
•
•
To assign a CC OUTPUT pattern/code to a particular sequence in the Setup>Sequences
menu:
1. Select the desired Sequence (or create a new Sequence).
2. In the Output CC field check the Enable box.
3. Press the Edit button to open the Sequence CC Outputs page.
4. Check the desired control output pin(s).
5. Press Done.
Activating, Editing, Deleting Sequences
To activate a sequence, in the Generate menu, press the Sequence button and select the
desired sequence. Sections 4.2.7.1 through 4.2.7.6 describe additional ways to activate.
Edit:
Edit Steps: Setup>Sequences>[select a sequence]>Edit Steps allows you to edit the steps
Delete:
74
12 separate ON/OFF switches assigned to separate calibration
sequences, or
a 12-bit wide bus allowing the user to define activation codes for up to
4095 separate calibration sequences.
active whenever a particular calibration sequence is operating, or
activated or deactivated as individual steps within a calibration
sequence.
Figure 3-26. Sequence CC Outputs Setup Page
Setup>Sequences>[select a sequence]>Edit allows you to edit a Sequence's
various attributes, such as name, timer, repeat count, input and output CCs.
with the selected Sequence.
Setup>Sequences>[select a sequence]>Delete will entirely delete the selected
Sequence.