5. Navigate to the Utilities>Diagnostics>Pressure Cal menu and select Photo
Press Cal.
6. In the Actual Pressure field input the value obtained in Step 3.
7. Once entered, press the Calibrate button and confirm in the next window to
proceed with the pressure calibration.
8. Wait for the Status field to show "Success" and then turn off the calibrator.
9. Reconnect power to the internal pump
10. Close the calibrator's panels.
A
I
6.5.
NALOG
NPUTS
Note
The Utilities>Diagnostics>Analog Inputs menu is used to calibrate the instrument's A-
to-D conversion circuitry. such as replacement of CPU, motherboard or power supplies.
083730750B DCN8606
Figure 6-14. In-Hg Pressure Cal for O
C
ALIBRATION
This calibration is typically only necessary after a major repair.

Figure 6-15. Analog Inputs Calibration Page

Photometer Option
3
117