Setting up Control Inputs for a Sequence

The calibrator's control inputs (Figure 2-3) allow the entire sequence to be triggered from
an external source. This feature allows the calibrator to operate in a slave mode so that
external control sources, such as a data logger can initiate the calibration sequences.
The calibrator's control inputs connector has:
To assign a CC INPUT pattern/code to a particular sequence in the Setup>Sequences
menu:
1. Select the desired Sequence (or create a new Sequence).
2. In the Input CC field check the Enable box.
3. Press the Edit button.
4. Check the desired control input pin(s).
5. Press Done.
083730750B DCN8606
Enable1 is for Status Block 1, which corresponds to the physical
CONTROL OUTPUT connector. To configure, press the respective
Enable box and check the desired box(es) whose numbers
correspond to the Pin numbers.
Enable 2 is for Status Block 2, which does not correspond to any
physical output but is used to communicate status over the serial
data port.
12 separate ON/OFF switches assigned to separate calibration
sequences or;
A 12-bit wide bus allowing the user to define activation codes for up to
4095 separate calibration sequences.