TCP Port1 allows choosing whether or not to display the command prompt, and to edit the
Port 1 number for defining the terminal control port by which terminal emulation software
addresses the instrument; also allows or disabling security on this port.

TCP Port2

TCP Port2 is configured with the port number for MODBUS.

Network Settings

The Setup>Comm>Network Settings menu is for Ethernet configuration. The address
settings default to automatic configuration by Dynamic Host Configuration Protocol
(DHCP). Most users will want to configure the instrument with a static IP address: click
the Static radio button to manually assign a static IP address (consult your network
administrator, and see Table 3-6 for information).
083730750B DCN8606
Enable/disable multidrop mode for multi-instrument configuration on a single
communications channel. Multidrop requires a unique ID for each instrument in
the chain (Setup>Vars>Instrument ID).
Select odd, or even, or no parity (typically set in conjunction with Data Bits and
Stop Bits).
Select among the communications protocols: TAPI (serial communication),
MODBUS RTU, or MODBUS ASCII. (See Section 5 for remote operation).
Enable/disable Quiet mode, which suppresses any feedback from the analyzer
(such as warning messages) to the remote device and is typically used when the
port is communicating with a computer program where such intermittent
messages might cause communication problems.
Such feedback is still available, but a command must be issued to receive them.
Enable/disable the rear panel COM2 Port for RS-485 communication. RS-485
mode has precedence over Multidrop mode if both are enabled. Also, RS-485
configuration disables the rear panel USB port.
Enable/disable the requirement for a password for this serial port to respond. The
only command that is active is the request-for-help command (? CR).
Select either 0 or 1 stop bit (typically set in conjunction with Parity and Data bits).