Table 8-7. DC Power Supply Acceptable Levels

POWER
SUPPLY
VOLTAGE
ASSY
PS1
+5
PS1
+15
PS1
-15
PS1
Agnd
PS1
Chassis
PS2
+12
PS2
Dgnd
I
C B
2
US
Operation of the I
distribution board & D2 on the Valve Driver PCA. Assuming that the DC power supplies
are operating properly, the I
board and D2 of the Valve Driver PCA are flashing
There is a problem with the I
the Valve Driver PCA are ON/OFF constantly.
T
OUCHSCREEN
Verify the functioning of the touch screen by observing the display when pressing a touch-
screen control button. Assuming that there are no wiring problems and that the DC power
supplies are operating properly, but pressing a control button on the touch screen does not
change the display, any of the following may be the problem:
You can verify this failure by logging on to the instrument using NumaView™ Remote
software or a terminal program. If the instrument responds to remote commands and the
display changes accordingly, the touch-screen interface may be faulty.
142
CHECK POWER DISTRIBUTION BOARD
TEST POINTS
FROM TEST POINT
NAME
#
Dgnd
1
Agnd
3
Agnd
3
Agnd
3
Dgnd
1
+12V Ret
6
+12V Ret
6
2
C bus can be verified by observing the behavior of D1 on the power
2
C bus is operating properly if D1 on the power distribution
2
C bus if both D1 on the power distribution board and D2 of
I
NTERFACE
The touch-screen controller may be malfunctioning.
The internal USB bus may be malfunctioning.
MIN V
TO TEST POINT
NAME
#
+5
2
4.8
+15
4
13.5
-15V
5
-14V
Dgnd
1
-0.05
Chassis
N/A
-0.05
+12V
7
11.75
Dgnd
1
-0.05
MAX V
5.25
16V
-16V
0.05
0.05
12.5
0.05
083730750B DCN8606