Test Channel Output; Table 4-21: Test Parameters Available For Analog Output A4 - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Operating Instructions

4.13.9. Test Channel Output

SAMPLE*
RANGE = 500.000 PPM
< TST TST > CAL
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
ALRM
SETUP X.X
ENTER DIAG PASS: 818
8
1
8
DIAG
SIGNAL I / O
NEXT
DIAG
ANALOG OUTPUT
PREV
NEXT

Table 4-21: Test Parameters Available for Analog Output A4

TEST CHANNEL
* This refers to the internal voltage level of the function NOT the output signal level
of the Test channel itself.
150
When activated, output channel A4 can be used to report one of the test functions
viewable from the SAMPLE mode display. To activate the A4 channel and select
a test function, follow this button sequence:
CO2 =X.XXX
SETUP
MORE
EXIT
EXIT
ENTR
EXIT
ENTR EXIT
ENTR EXIT
NONE
CO
MEASURE
2
CO
REFERENCE
2
SAMPLE PRESS
SAMPLE FLOW
SAMPLE TEMP
BENCH TEMP
WHEEL TEMP
CHASSIS TEMP
PHT DRIVE
Teledyne API Model T360/T360M Operation Manual
Continue to press NEXT until ...
DIAG
TEST CHAN OUTPUT
PREV
NEXT
DIAG TCHN
TEST CHANNEL: NONE
NEXT
DIAG TCHN
TEST CHANNEL: CO2 MEASURE
PREV
NEXT
Press PREV or NEXT
Press ENTR to select
to move through the
the displayed
list of available
parameter activating
parameters
the test channel.
(Table 6-13)
ZERO
FULL SCALE
Test Channel is turned off
0 mV
5000 mV*
0 mV
5000 mV*
0 "Hg
0 cc/m
1000 cc/m
0C
0C
0C
0C
0 mV
ENTR
EXIT
ENTR
EXIT
ENTR
EXIT
Press EXIT to
return to the
DIAG menu
40 "Hg
70C
70C
70C
70C
5000 mV
07272B DCN6552