Teledyne T360 Operation Manual page 178

Carbon dioxide analyzer
Table of Contents

Advertisement

Calibration Procedures
The CO
span
2
concentration values
automatically default to
400.0 Conc .
To change this value to
the actual concentration of
the span gas, enter the
number by pressing the
button under each digit
until the expected value
appears.
For this Initial Calibration it is important to independently verify the PRECISE CO
If the source of the Span Gas is from a Calibrated Bottle, use the exact concentration value
176
STEP TWO: Set the expected CO
SAMPLE*
RANGE = 500.000 PPM
< TST TST >
CAL
M-P CAL
RANGE = 500.000 PPM
< TST TST >
ZERO
M-P CAL
CO2 SPAN CONC: 450.000 Conc
0
0
0
Value of the SPAN gas.
printed on the bottle.
STEP THREE: Perform the zero/span calibration. Zero and span checks using
the zero/span valve option are similar to that described in Section 5.5, except that
zero air and span gas is supplied to the analyzer through the zero/span valves
rather than through the sample inlet port.
The zero and cal operations are initiated directly and independently with
dedicated buttons (CALZ & CALS).
Teledyne API Model T360/T360M Operation Manual
Span Gas concentration:
2
CO2 =X.XXX
CALZ CALS
SETUP
CO2 =X.XXX
CONC
EXIT
4
5
.0
ENTR EXIT
NOTE
This sequence causes the
analyzer to prompt for the
expected CO
span
2
concentration.
EXIT ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display.
Concentration
2
07272B DCN6552