Teledyne API Model T360/T360M Operation Manual
Question
How long does the IR source
last?
Where is the sintered
filter/sample flow control orifice?
How do I set up a SEQUENCE to
run a nightly calibration check?
How do I set the analog output
signal range and offset?
What is the averaging time for a
T360?
07272B DCN6552
Answer
Typical lifetime is about 2-3 years.
These components are located inside the flow control assembly that
is attached to the inlet side of the sample pump, see Figure 3-5.
See Section 8.6.1 for instructions on disassembly and replacement.
The setup of this option is located in Section 5.6.
Instructions for this can be found in Section 4.13.4 which describes
analog I/O configuration.
The default averaging time, optimized for ambient pollution
monitoring, is 150 seconds for stable concentrations and 10
seconds for rapidly changing concentrations; see Section 7.6.1 for
more information. However, it is adjustable over a range of 0.5
second to 200 seconds (please contact Technical Support for more
information).
Troubleshooting and Service
261