Initial Calibration Procedure - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Getting Started

3.8. INITIAL CALIBRATION PROCEDURE

Press this button to set
the analyzer for SNGL
DUAL or AUTO ranges
To change the value of the
reporting range span, enter the
number by pressing the key under
each digit until the expected value
appears.
58
The next task is to calibrate the analyzer.
To perform the following calibration you must have sources for zero air and span
gas available for input into the sample port on the back of the analyzer. See
Section 3.6 for instructions for connecting these gas sources.
While it is possible to perform this procedure with any range setting we
recommend that you perform this initial checkout using the 500 ppm range.
The following procedure assumes that the instrument does not have any of
the available Zero/Span Valve Options installed. See Section 5.4 for
instructions for calibrating instruments with Z/S valve options.
1. Set the Analog Output Range:
SAMPLE
RANGE = 500.000 PPM
< TST TST > CAL
SETUP X.X
CFG DAS
RNGE
PASS CLK MORE
SETUP X.X
RANGE CONTROL MENU
MODE SET UNIT
SETUP X.X
RANGE: 500.000 CONC
0
0
5
0
SETUP X.X
RANGE:
0
0
0
5
Teledyne API Model T360/T360M Operation Manual
CO2
=X.XXX
SETUP
EXIT
EXIT
0
.0
ENTR EXIT
500.000 Conc
0
.0
ENTR EXIT
Press this button to select the
concentration units of measure:
PPB, PPM, UGM, MGM
EXIT ignores the new setting and
returns to the RANGE CONTROL
MENU.
ENTR accepts the new setting and
returns to the
RANGE CONTROL MENU.