Teledyne API Model T360/T360M Operation Manual
MODE AND ATTRIBUTE
Sequence
Mode
Timer Enable
Starting Date
Starting Time
Delta Days
Delta Time
Duration
Calibrate
The programmed STARTING_TIME must be a minimum of 5 minutes later tha The real time
clock (see Section 4.10 for setting real time clock).
Avoid setting two or more sequences at the same time of the day. Any new sequence which is
initiated whether from a timer, the COM ports, or the contact closure inputs will override any
sequence which is in progress.
If at any time an illegal entry is selected (Example: Delta Days > 367) the ENTR button will
disappear from the display.
07272B DCN6552
The following example sets Sequence #2 to do a Zero-Span Calibration every
other day starting at 1 am on September 4, 2001, lasting 15 minutes, without
calibration. This will start ½ hour later each iteration.
VALUE
2
ZERO-SPAN
ON
Sept. 4, 2001
01:00
2
00:30
15.0
NO
Define Sequence #2
Select Zero and Span Mode
Enable the timer
Start after Sept 4, 2001
First Span starts at 1:00AM
Do Sequence #2 every other day
Do Sequence #2 ½ hr later each day
Operate Span valve for 15 min
Do not calibrate at end of Sequence
NOTES
Calibration Procedures
COMMENT
181