Teledyne T360 Operation Manual page 65

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
3.8.1.2. O
CALIBRATION METHOD
2
SAMPLE
RANGE = 500.000 PPM
< TST TST >
CAL
SAMPLE
GAS TO CAL:CO2
CO2
O2
SAMPLE
GAS TO CAL:O2
NOX
O2
The O
span concentration value automatically defaults to
2
If this is not the the concentration of the span gas being
used, toggle these buttons to set the correct concentration
of the O
07272B DCN6552
STEP 1 – SET O
SPAN GAS CONCENTRATION :
2
Set the expected O
span gas concentration.
2
This should be equal to the percent concentration of the O
selected reporting range (default factory setting = 20.8%; the approximate O
content of ambient air).
CO2 =XXX.X
SETUP
ENTR EXIT
M-P CAL
<TST TST> ZERO SPAN
ENTR
EXIT
M-P CAL
20.8 %.
calibration gases.
2
A1:NXCNC1 =100PPM
CONC
O2 SPAN CONC:20.8%
0
2
0
.8
0
Getting Started
span gas of the
2
NOX=X.XXX
EXIT
EXIT ignores the new
ENTR EXIT
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the previous menu.