Table 4-30: Default Hessen Status Bit Assignments - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Operating Instructions
4.15.4.7. SETTING HESSEN PROTOCOL STATUS FLAGS
It is possible to assign more than one flag to the same Hessen status bit. This allows the
grouping of similar flags, such as all temperature warnings, under the same status bit.
Be careful not to assign conflicting flags to the same bit as each status bit will be triggered if any
of the assigned flags is active.
166
Teledyne API's implementation of Hessen protocols includes a set of status bits
that the instrument includes in responses to inform the host computer of its
condition. Each bit can be assigned to one operational and warning message flag.
The default settings for these bit/flags are:

Table 4-30: Default Hessen Status Bit Assignments

STATUS FLAG NAME
WARNING FLAGS
SAMPLE FLOW WARNING
BENCH TEMP WARNING
SOURCE WARNING
BOX TEMP WARNING
WHEEL TEMP WARNING
SAMPLE TEMP WARNING
SAMPLE PRESSURE WARNING
INVALID CONC
(The Instrument's Front Panel Display Will Show The
Concentration As "XXXX")
OPERATIONAL FLAGS
Instrument Off
In Manual Calibration Mode
In Zero Calibration Mode
In Span Calibration Mode
UNITS OF MEASURE FLAGS
UGM
MGM
PPB
PPM
SPARE/UNUSED BITS
UNASSIGNED FLAGS (0000)
Sync Warning
Conc Alarm 1
Conc Alarm 2
Photo Temp Warning
System Reset
Rear Board Not Detected
Teledyne API Model T360/T360M Operation Manual
Relay Board Warning
Analog Cal Warning
Cannot Dyn Zero
Cannot Dyn Span
Invalid Conc
NOTES:
DEFAULT BIT
ASSIGNMENT
0001
0002
0004
0008
0010
0020
0040
0080
0100
0200
0400
0800
0000
2000
4000
6000
100, 1000, 8000