Figure 7-7: Oxygen Sensor - Principle Of Operation - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
7.2.1.1. O
SENSOR OPERATION WITHIN THE T360 ANALYZER
2
07272B DCN6552
Figure 7-7:
Oxygen Sensor - Principle of Operation
The oxygen sensor option is transparently integrated into the core analyzer
operation. All functions can be viewed or accessed through the front panel, just
like the functions for CO
The O
concentration is displayed in the upper right area of the display with
2
CO
concentration.
2
Test functions for O
with the analyzer's other test functions.
O
sensor calibration is performed via the front panel CAL function and is
2
performed in a nearly identical manner as the standard CO
Section 3.8.1 for more details.
Stability of the O
sensor can be viewed via the front panel display.
2
A signal representing the currently measured O
The O
concentration range is 0-100% (user selectable) with 0.1% precision and
2
accuracy and is available to be output via the instrument's analog output channel
A3 (see Section 4.13.4).
The temperature of the O
PID loop and can be viewed on the front panel as test function O2 TEMP.
The O
sensor assembly itself does not have any serviceable parts and is enclosed
2
in an insulated canister.
2
slope and offset are viewable from the front panel along
2
sensor is maintained at a constant 50° C by means of a
2
Principles of Operation
calibration. See
2
concentration is available .
2