Teledyne T360 Operation Manual page 300

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API - T360/T360M, 360E/360EM Appendix A Menu Trees (05233C DCN6552)
Setup Variable
REF_SDEV_LIMIT
REF_SOURCE_LIMIT
FACTORY_OPT
A-22
Numeric
Default
Units
Value
mV
50
mV
3000 (not
used)
Warnings:
1100–4800,
25–4800
BitFlag
512,
5
768
Value Range
0.1–500
1–5000
3, 4, 15
0–0x7fffffff
Description
Reference detector standard
deviation must be below this limit
to switch out of startup mode.
Reference source warning limits.
Set point is not used.
Factory option flags. Add values
to combine flags.
1 = enable dilution factor
2 = zero/span valves installed
4 = enable conc. alarms
8 = enable linearity adjustment
factor
16 = display units in
concentration field
32 = enable software-controlled
maintenance mode
3, 5
64
= span valve installed
128 = enable switch-controlled
maintenance mode
256 = compute only offset during
zero calibration
512 = 220 V A/C power
1024 = non-zero offset
calibration (linearity adjustment
must also be enabled)
2048 = enable Internet option
4096 = use "old" style numeric
data entry menus when editing
conc. table
8192 = locate high range and
zero cal. status outputs on relays
16384 = enable external analog
25
inputs
07272B DCN6552