Figure 5-2: Pneumatic Connections-Basic Configuration-Using Gas Dilution Calibrator - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
Figure 5-2:
Pneumatic Connections–Basic Configuration–Using Gas Dilution Calibrator
The CO
span
2
concentration values
automatically default to
400.0 Conc.
To change this value to
the actual concentration of
the span gas, enter the
number by pressing the
button under each digit
until the expected value
appears.
For this Initial Calibration it is important to independently verify the PRECISE CO
Value of the SPAN gas.
If the source of the Span Gas is from a Calibrated Bottle, use the exact concentration value
printed on the bottle.
07272B DCN6552
STEP TWO: Set the expected CO
SAMPLE*
RANGE = 500.000 PPM
< TST TST >
CAL
M-P CAL
RANGE = 500.000 PPM
< TST TST >
ZERO
M-P CAL
CO2 SPAN CONC: 400.000 Conc
0
0
0
Span Gas concentration:
2
CO2 =X.XXX
SETUP
CO2 =X.XXX
CONC
EXIT
4
5
.0
ENTR EXIT
NOTE
Calibration Procedures
This sequence causes the
analyzer to prompt for the
expected CO
span
2
concentration.
EXIT ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display..
Concentration
2
171