Teledyne T360 Operation Manual page 139

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
4.13.4.1. ANALOG OUTPUT SIGNAL TYPE AND RANGE SPAN SELECTION
DIAG
PREV
DIAG AIO
DIAG AIO
DIAG AIO
DIAG AIO OUTPUT RANGE:
To set the signal
level and type of
0.1V
the selected
channel
DIAG AIO OUTPUT RANGE:
0.1V 1V
07272B DCN6552
To select an output signal type (DC Voltage or current) and level for one output
channel, activate the ANALOG I/O CONFIGURATION MENU (see Section
4.13.4) then press:
FROM ANALOG I/O CONFIGURATION MENU
ANALOG I / O CONFIGURATION
NEXT
AOUTS CALIBRATED: NO
< SET
SET>
CAL
CONC_OUT_2:5V, CAL
< SET SET>
EDIT
CONC_OUT_2 RANGE: 5V
SET>
EDIT
5V
1V
5V
10V
CURR
10V
5V
10V CURR
ENTR
EXIT
EXIT
EXIT
EXIT
ENTR
EXIT
ENTR EXIT
Operating Instructions
Press SET> to select the
analog output channel to be
configured. Press EDIT to
continue
Pressing ENTR records the new setting
and returns to the previous menu.
Pressing EXIT ignores the new setting
and returns to the previous menu.
137