Operating Instructions
4.13.4.2. ANALOG OUTPUT CALIBRATION MODE
Exit at any time
to return to the
main DIAG
menu
If any of the channels have
not been calibrated this
message will read NO.
138
The analog outputs can be calibrated automatically or manually. In its default
mode, the instrument is configured for automatic calibration of all channels.
Manual calibration should be used for the 0.1V range or in cases where the
outputs must be closely matched to the characteristics of the recording device.
Outputs configured for automatic calibration can be calibrated as a group or
individually. Calibration of the analog outputs needs to be carried out on first
startup of the analyzer (performed in the factory as part of the configuration
process) or whenever re-calibration is required.
To calibrate the outputs as a group, activate the ANALOG I/O
CONFIGURATION MENU (see Section 4.13.1), then press:
FROM DIAGNOSTIC MENU:
DIAG
ANALOG I / O CONFIGURATION
PREV
NEXT
DIAG AIO
AOUTS CALIBRATED:
< SET SET>
CAL
DIAG AIO
AUTO CALIBRATING CONC_OUT_1
AUTO CALIBRATING CONC_OUT_2
AUTO CALIBRATING TEST_OUTPUT
DIAG AIO
AOUTS CALIBRATED:
< SET SET> CAL
Teledyne API Model T360/T360M Operation Manual
ENTR
EXIT
NO
If AutoCal has been
EXIT
turned off for any
channel, the message
for that channel will be
NOT AUTO CAL
YES
EXIT
similar to:
CONC_OUT_1
Exit to return to
the I/O
configuration
menu
07272B DCN6552