4.8.5. Auto Range Mode (AUTO)

SAMPLE*
RANGE = 500.000 PPM
< TST TST > CAL
PRIMARY SETUP MENU
SETUP X.X
CFG DAS
RNGE
PASS CLK MORE
SETUP X.X
RANGE CONTROL MENU
MODE
SET UNIT
SETUP X.X
RANGE MODE:
SNGL DUAL
AUTO
In AUTO range mode the LOW and HIGH ranges have separate slopes and offsets for computing
CO2 concentration. The two ranges must be independently calibrated.
Avoid accidentally setting the low range of the instrument with a higher span limit tha The high
range. This will cause the unit to stay in the low reporting range perpetually and defeat the
100
In AUTO range mode, the analyzer automatically switches the reporting range
between two user-defined ranges (low and high). The unit will switch from low
range to high range when the CO
span. The unit will return from high range back to low range once both the CO
concentration falls below 75% of the low range span.
In AUTO Range mode the instrument reports the same data in the same range on
both the A1 and A2 outputs and automatically switches both outputs between
ranges as described above. Also, the RANGE test function displayed on the front
panel will automatically switch to show which range is in effect.
The high/low range status is also reported through the external, digital status
outputs (Section 4.15.1).
To set individual ranges press the following menu button sequence.
SETUP X.X
SNGL DUAL AUTO
CO2 =X.XXX
SETUP
SETUP X.X
MODE
EXIT
SETUP X.X
0
EXIT
SNGL
SETUP X.X
ENTR EXIT
0
function of the AUTO range mode.
Teledyne API Model T360/T360M Operation Manual
concentration exceeds 98% of the low range
2
RANGE MODE:
AUTO
RANGE CONTROL MENU
SET
UNIT
LOW RANGE:
500.0 Conc
0
5
0
0
.0
HIGH RANGE:
500.0 Conc
0
5
0
0
.0
NOTE
NOTE
ENTR
EXIT
EXIT x 2 returns
to the main
SAMPLE display
EXIT
Toggle the numeral
buttons to set the
ENTR EXIT
LOW and HIGH
range value.
ENTR accepts the
new setting, EXIT
ignores the new
setting.
ENTR EXIT