Teledyne T360 Operation Manual page 62

Carbon dioxide analyzer
Table of Contents

Advertisement

Getting Started
3. Perform the Zero/Span Calibration Procedure
The SPAN button now
appears during the
transition from zero to
span..
If either the ZERO or
SPAN buttons fail to
appear see Section 11
for troubleshooting tips.
60
SAMPLE*
RANGE = 500.000 PPM
< TST
TST
>
CAL
SAMPLE
STABIL=XXX.X PPM
< TST TST >
CAL
ACTION:
Allow zero gas to enter the sample port at the
rear of the instrument.
M-P CAL
STABIL=XXX.X PPM
< TST TST >
CAL
M-P CAL
STABIL=XXX.X PPM
< TST TST >
ZERO
CONC
M-P CAL
STABIL=XXX.X PPM
< TST TST >
ENTR
CONC
ACTION:
Allow span gas to enter the sample port at the
rear of the instrument.
M-P CAL
STABIL=XXX.X PPM
< TST TST >
SPAN
CONC
M-P CAL
RANGE = 500.000 PPM
< TST TST >
ENTR
SPAN CONC
M-P CAL
RANGE = 500.000 PPM
< TST TST > ENTR
CONC
Teledyne API Model T360/T360M Operation Manual
CO2 =XXX.X
SETUP
CO2 =XXX.X
SETUP
CO2 =XXX.X
SETUP
CO2 =XXX.X
EXIT
CO2 =XXX.X
EXIT
CO2 =XXX.X
EXIT
CO2 =XXX.X
EXIT
CO2 =XXX.X
EXIT
Set the Display to show the
STABIL test function.
This function calculates the
stability of the CO
x
measurement
Wait until STABIL
falls below 1.0 ppm.
This may take several
minutes.
Press ENTR to changes the
OFFSET & SLOPE values for the
CO
measurements.
2
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
The value of
STABIL may jump
significantly.
Wait until it falls back
below 1.0 ppm
This may take several
minutes.
Press ENTR to change the
OFFSET & SLOPE values for the
CO
measurements.
2
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
EXIT returns to the main
SAMPLE display
07272B DCN6552