Teledyne T360 Operation Manual page 179

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual
The SPAN button now
appears during the
transition from zero to
span.
If either the ZERO or
SPAN buttons fail to
appear see Section 11
for troubleshooting tips.
07272B DCN6552
SAMPLE*
RANGE = 500.000 PPM
<TST
TST>
CAL
CALZ
CALS SETUP
SAMPLE
STABIL=XXX.X PPM
< TST TST >
CAL
CALZ
CALS SETUP
ACTION:
Allow zero gas to enter the sample port at the
rear of the instrument.
M-P CAL
STABIL=XXX.X PPM
< TST
TST
>
CAL
CALZ
CALS
M-P CAL
STABIL=XXX.X PPM
< TST TST >
ZERO
CONC
M-P CAL
STABIL=XXX.X PPM
< TST TST >
ENTR
CONC
ACTION:
Allow span gas to enter the sample port at the
rear of the instrument.
M-P CAL
STABIL=XXX.X PPM
< TST TST >
SPAN
CONC
M-P CAL
RANGE = 500.000 PPM
< TST TST >
ENTR
SPAN CONC
M-P CAL
RANGE = 500.000 PPM
< TST TST > ENTR
CONC
CO2 =XXX.X
Set the Display to show the
This function calculates the
CO2 =XXX.X
CO2 =XXX.X
SETUP
CO2 =XXX.X
EXIT
Press ENTR to changes the
CO2 =XXX.X
OFFSET & SLOPE values for the
EXIT
Press EXIT to leave the calibration
unchanged and return to the
CO2 =XXX.X
EXIT
Press ENTR to change the
OFFSET & SLOPE values for the
CO2 =XXX.X
Press EXIT to leave the calibration
EXIT
unchanged and return to the
CO2 =XXX.X
EXIT returns to the main
SAMPLE display
EXIT
Calibration Procedures
STABIL test function.
stability of the CO
x
measurement
Wait until STABIL
falls below1.0 ppm.
This may take several
minutes.
CO
measurements.
2
previous menu.
The value of
STABIL may jump
significantly.
Wait until it falls back
below 1.0 ppm.
This may take several
minutes.
CO
measurements.
2
previous menu.
177