Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
4.2.3. Startup / Warm Up of the M501-TRS
Figure 4-6. M501-TRS Temperature Controller Startup
07267B DCN6485
After electrical and pneumatic connections are made, turn on the instrument and pump
power. The exhaust fan should start.
The M501-TRS' temperature controller is preprogrammed at the factory so no special
setup operation is required.
immediately come on in operation mode: the current temperature of the converter oven
should the display immediately appear in the display area and the process value (PV)
LED should be lit.
It may take as much as 30 minutes for the oven to reach its nominal operating
temperature.
During that initial warm up period the high and low alarms and the M501-TRS single
alarm output are disabled. Both the internal alarms and the alarm output will be
automatically enabled once the converter oven temperature rises above the lower alarm
limit.
The temperature controller (see Figure 4-6) should
Pneumatic Connections
31