To Display The Current Temperature; To Manually Adjust The Converter Oven Temperature - Teledyne T102 Operation Manual

Total reduced sulfur analyzer with thermal converter
Table of Contents

Advertisement

Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual

7.2. To Display The Current Temperature

7.3. To Manually Adjust the Converter Oven Temperature

Press the PV/SV mode key
Press the 10's UP key once
Press the DOWN key
Press the 100's UP Key once
Press and hold the 100's UP key
--
Press the ENTER key
Press the PV/SV mode key
07267B DCN6485
If the Process Value is displaying a numerical value, the process value is currently
being displayed.
If it is not being displayed press the select key for 2 seconds.
DO NOT SET THE TEMPERATURE HIGHER THAN 1050 ºC
4. Set the main display to show the current value of the set variable by
pressing the PV/SV mode key.
5. To set each digit:
c.
Press the up-arrow under that digit once. The digit will flash.
d.
To increment that digit press and hold the UP key until the
appropriate number is displayed.
e.
To decrement that digit press and hold the DOWN key until the
appropriate number is displayed.
f.
To increment/decrement the 1000's digit it is necessary to adjust
increment/decrement the 100's digit up and down. Each time the
100's digit passes "0" the 1000s digit will increment or decrement
correspondingly. See EXAMPLE table below.
6. Once the desired value is reached, press the ENT key to store the new
set value
7. Return the main display to process mode by pressing the PV/SV mode
button once.
EXAMPLE to change the set value from 950 to equal 1050.
ACTION
M501-TRS OPERATING INSTRUCTIONS
CAUTION !
RESULT
The SV indicator will lit up and the display will
show 950.
The 10's digit will begin to blink
The 10's digit will decrement from "5". Release
the DOWN key when the 10's digit reads "5".
The 100's digit will begin to blink
The 100's digit will increment from "9". When it
passes "0" the 1000's digit will increment to "1".
Release the 100's UP key.
The Display should now read "1050"
The new set value is recorded
The current level of the process value will be
displayed.