Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual

11.6.1. Temperature Controller Primary Menu Parameters

Tables 10-3 and 10-4 show the typical primary P-I-D parameter values for operation
The P, I and d values may be different for other AC main voltages (call TAPI Technical
Support), and will vary somewhat after auto-tuning.
07267B DCN6485
on 115V/60Hz with a set value of 850°C.
To set these parameters:
1. Open the appropriate menu
a. Press the SELECT key once to access the primary menu parameters.
The main display will show
b. Press and Hold the SELECT key for 3 seconds to access the primary
menu parameters. The main display will show
2. Use the SELECT, DOWN or 100's UP key to scroll through the primary
menu parameters until the appropriate parameter is displayed (see Table
10-3 or Table 10-4).
3. Press the DATA key once. The current value of the parameter will be
displayed.
4. To set each digit:
a. Press the up-arrow under that digit once. The digit will flash.
b. To increment that digit, press and hold the digit until the appropriate
number is displayed.
c. To decrement that digit press and hold the DOWN key until the
appropriate number is displayed.
d. To increment/decrement the 1000's digit it is necessary to adjust
increment/decrement the 100's digit up and down. Each time the
100's digit passes "0" the 1000s digit will increment or decrement
correspondingly.
5. Once the desired value is reached, press the ENT key to store the new
set Parameter value.
6. Press the PV/SV mode key to return to operational mode.
NOTE
.
TROUBLESHOOTING and REPAIR
.