TROUBLESHOOTING and REPAIR
11.5. Additional Repair Procedures
70
10. If TRS SO
converter efficiency is below 90% check for one of the
2
following possible causes:
Pneumatic leak
•
Perform a leak check (with the M501-TRS connected) as
•
described in the Detailed Pressure Leak Check section in the
T101 Manual - PN 07266.
Plugged or constricted pneumatic flow
•
•
Perform a sample flow check as described in the Performing a
Sample Flow Check section in the T101 Manual - PN 07266. This
tests the entire system
Disconnect the gas feed line from the FROM ANALYZER port of
•
the M501-TRS. Attach the flow meter to the port and repeat the
test. Isolates the portions of the pneumatic system down stream
from the TRS SO
in the flow test.
Attach the flow meter to the FROM CONVERTER port at the
•
back of the T102. Retest. This bypasses M501-TRS during the
flow test.
Improper Set point temperature
•
•
Call Teledyne API's Technical Support for advice.
Incorrect/wrong concentration Span Gas used in efficiency
•
test
•
Independently verify the concentration of the span gas.
Contaminants in the converter chamber, pneumatic lines
•
Sometimes contaminants such as tiny particles of scrubber
•
material from the M501-TRS' SO
converter tube and react with the TRS SO
process. This can occur if the filter material inside the scrubber
lets any particulates through.
Call TAPI Technical Support for instructions on cleaning the
•
converter tube and M501-TRS pneumatic lines.
The following repair procedures are in addition to those listed in the Repair Procedures
section in the T101 Manual - PN 07266,
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
switching valve but includes the M501-TRS
2
scrubber can get into the
2
conversion
2