M501-TRS OPERATING INSTRUCTIONS

7.4. Autotune the Temperature Controller

Before initiating the autotune feature make sure that the converter temperature oven

7.4.1. Initiating the Autotune Process

The P-I-D parameters calculated by autotuning will be retained even if the power is
lost. However, if the power is turned off during the auto-tuning process, you must
46
The M501-TRS controller includes an auto tune feature which allows the controller to
find and set optimum values for various process control parameters so that the controller
can establish and maintain the converter oven at the temperature set value in the most
stable and efficient manner.
has reached a stable, constant temperature.
1. Press the SELECT key once. The main display will show
2. Use the SELECT , DOWN or 100's UP key to scroll through the primary
menu parameters until the display shows
3. Press the DATA key once. The display will show
4. Press the 1's UP key once. The display will show
on set point value).
5. Press the ENTER key to begin the autotune process. A blinking decimal
point will appear at the bottom right-hand corner of the main display.
6. Wait until the blinking light stops. This may take up to 30 minutes.
7. The autotune process is finished. The autotune parameter value will
automatically reset to zero (off).
8. Press the PV/SV mode key to return to operational mode.
restart autotuning.
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
NOTE
Note
(AT =Autotune).
(zero = Off).
(1 = autotune based
07267B DCN6485