TROUBLESHOOTING and REPAIR
11.6. Manually Programming the M501-TRS Temperature
Controller
The temperature controller has been programmed at the factory and should not be
DO NOT manually alter the PID parameters of the M501-TRS temperature Controller
unless directed to do so by TAPI Technical Support personnel.
84
verify that all connections are leak free. If a leak is detected, the leak
should be resolved before continuing.
20. Install the inner cover of the Heater Block and secure with the three
nuts. Close the Front Panel and secure with the four nuts. Install the
top cover on the Converter chassis.
21. The Converter is now ready for the application of power. You will be
looking for an indication from the temperature controller that it is
functioning correctly and driving the heater to the desired "set"
temperature. Apply power now.
22. Check the Temp Controller to be sure that it knows which type of
Thermocouple it has in it. Follow the directions in Section 7.3 of this
manual to be sure that the temp controller is set properly.
23. After the Converter comes to the regulated temperature, perform the
Auto Tune function (see Section 7.4) to tune the Temperature Controller
to the new thermocouple. .
24. After the Auto-Tune process is completed, verify that the "process"
temperature is indicating that the desired temperature is stable and
being regulated.
25. The converter is now ready for operation.
altered, (except for temperature set point).
In the event that the temperature controller of the M501-TRS must be replaced, some
initial programming is required to set the control functions if reprogramming is
necessary. The tables in the following section define the approximate initial values for
various processes control parameters. Once they are set, perform an autotune procedure
as defined in Section 7.4 of this addendum
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
NOTE
07267B DCN6485