SPECIFICATIONS AND APPROVALS

2.3. CE Mark Compliance

16
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
Range: Any range from 50 parts per billion (ppb) to 10 parts per million
(ppm).
Ambient temperature range of 5
Line voltage range of 105-125 VAC or 220-240 VAC, at 50 or 60 Hz.
Sample filter: Equipped with PTFE filter element in the internal filter
assembly.
Sample flow of 650 +/- 65 cc/min.
Vacuum pump (internal or external) capable of 14"Hg Absolute pressure
@ 1 slpm or better.
Software settings:
Dynamic span
Dynamic zero
Dilution factor
AutoCal
Dual range
Auto-range
Temp/Pressure compensation
Under the designation, the analyzer may be operated with or without the following
optional equipment:
Rack mount with chassis slides.
Rack mount without slides, ears only.
Zero/span valve options.
Internal zero/span (IZS) option with either:
SO2 permeation tube - 0.4 ppm at 0.7 liter per minute;
certified/uncertified.
SO2 permeation tube - 0.8 ppm at 0.7 liter per minute;
certified/uncertified. Under the designation, the IZS option cannot be
used as the source of calibration.
4-20mA isolated analog outputs.
Status outputs.
Control inputs.
RS-232 output.
Ethernet output.
Zero air scrubber.
4-20mA, isolated output.
See Section titled CE Mark Compliance in the T101 Manual - PN 07266
o
o
C to 40
C.
OFF
OFF
OFF
ON or OFF
ON or OFF
ON or OFF
ON
07267B DCN6485