Attach a gas line from the source of zero air (e.g. a Teledyne API M701 zero air
Generator) to the ZERO AIR inlet.
Zero air can be supplied by the API M701 zero air generator.
A restrictor is required to regulate the gas flow at 2x the gas flow of the analyzer.
V
ENTING
In order to prevent back diffusion and pressure effects, both the span gas and zero air
supply lines should be:
Vented outside the enclosure
Not less than 2 meters in length
Not greater than 10 meters in length
E
XHAUST
Attach an exhaust line to the EXHAUST outlet fitting. The exhaust line should be:
¼" PTFE tubing
A maximum of 10 meters long
Vented outside the T400 analyzer's enclosure
Once the appropriate pneumatic connections have been made, check all pneumatic
fittings for leaks using the procedures defined in Section 5.6.5).
42
O
UTLET
CAUTION – Ozone Exposure Risk
Venting should be outside the shelter or immediate area surrounding
the instrument and conform to all safety requirements regarding
exposure to O
.
3
Teledyne API – Model T400 O
Analyzer
083730400A DCN8130
3