8. Locate the potentiometer used to adjust the O
generator UV output.
3
O
Generator
3
Body
Adjustment
Pot
O
3
Generator
Reference
Detector
PCA

Figure 5-10. Location of O

Generator Adjustment Pot
3
9. Navigate to Utilities>Diagnostics>Analog Inputs>O3 Gen Ref.
10. Using an insulated pot adjustment tool, turn the O
Generator Adjustment Pot
3
(Figure 5-10) until the value of O3 Gen Ref is 2500.0 mV ±100 mV.
11. Finish screwing in the UV lamp holder so that it is snug but not excessively tight.
12. Replace the analyzer's cover.
13. Perform a leak check (Section 5.6.5).
14. Perform an ozone generator calibration (Section 4.5.3).