For the Model T400 photometric ozone analyzer in its basic configuration (i.e. without
the optional internal zero air source or valves), attach the following pneumatic lines:
S
G
AMPLE
Attach a sample inlet line to the SAMPLE inlet port.
Sample Gas pressure must equal ambient atmospheric pressure (1.0 psig)
In applications where the sample gas is received from a pressurized manifold, a vent must
be placed on the sample gas line. This vent line must be:
• At least 0.2m long
• No more than 2m long
• Vented outside the shelter or immediate area surrounding the instrument
C
ALIBRATION
The source of calibration gas is also attached to the SAMPLE inlet, but only when a
calibration operation is actually being performed.
E
XHAUST
Attach an exhaust line to the analyzer's EXHAUST outlet fitting. The exhaust line
should be:
PTFE tubing; minimum O.D ¼";
A maximum of 10 meters long;
Vented outside the analyzer's enclosure.
L
C
EAK
HECK
Once the appropriate pneumatic connections have been made, check all pneumatic
fittings for leaks using the procedures defined in Section 5.6.3.
40
S
AS
OURCE
G
S
AS
OURCES
O
UTLET
Teledyne API – Model T400 O
Analyzer
083730400A DCN8130
3