2.3.1.6 CONNECTING THE CONCENTRATION ALARM RELAY (OPTION 61)
The concentration alarm option provides four (4) "dry contact" relays on the rear panel
(Figure 2-8) , each with 3 pins: Common (C), Normally Open (NO), and Normally
Closed (NC). The Relays can be mapped to reflect various internal instrument conditions
and states.
(Section 2.5.7) under MB Relay [1 thru 4].
2.3.1.7 CONNECTING THE COMMUNICATION INTERFACES
For remote communications, the rear panel provides Ethernet, USB, RS-232, optional
RS-232 Multidrop, and optional RS-485 connectors. In addition to using the appropriate
cables, each type of communication method must be configured using the
SETUP>COMM menu, Section 2.5.10).
E
THERNET
For network or Internet communication with the analyzer, connect an Ethernet cable from
the analyzer's rear panel Ethernet interface connector to an Ethernet access port.
Although Ethernet is DHCP-enabled by default, it should be manually configured
(Section 2.5.10.5) with a static IP address.
USB (O
PTION
The rear panel USB option is for direct communication between the analyzer and a PC.
Connect a USB cable between the analyzer and computer USB ports. The baud rate for
the analyzer and the computer must match; see Section 3.2.
Note
083730400A DCN8130
ConFigure these outputs through the Setup>Digital Outputs menu
Figure 2-8. Concentration Alarm Relay
C
ONNECTION
) C
ONNECTION
If this option is installed, the COM2 port cannot be used for
anything other than Multidrop communication.
Teledyne API – Model T400 O
Analyzer