Default Das Channels; Table 7-2: Das Data Channel Properties - Teledyne T802 Operation Manual

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

Section II Operating Instructions

Table 7-2: DAS Data Channel Properties

PROPERTY
NAME
The name of the data channel.
TRIGGERING
The event that triggers the data channel to measure
EVENT
and store the datum
NUMBER AND
A User-configurable list of data types to be
LIST OF
recorded in any given channel.
PARAMETERS
The amount of time between each channel data
REPORT PERIOD
point.
The number of reports that will be stored in the data
NUMBER OF
file. Once the limit is exceeded, the oldest data is
RECORDS
over-written.
Enables the analyzer to automatically report
RS-232 REPORT
channel values to the RS-232 ports.
CHANNEL
Enables or disables the channel. Allows a channel
ENABLED
to be temporarily turned off without deleting it.
Disables sampling of data parameters while
CAL HOLD OFF
instrument is in calibration mode.
1
More with APICOM, but only the first six are displayed on the front panel).
2
For information regarding the maximum number of channels, parameters, and records and how to calculate the file size for each data
channel, refer to the DAS manual downloadable from the TAPI website at http://www.teledyne-api.com/manuals/ under Special
Manuals.
3
When enabled records are not recorded until the DAS HOLDOFF period is passed after calibration mode. DAS HOLDOFF SET in the
5.8
VARS menu (see Section

7.1.2. DEFAULT DAS CHANNELS

140
Teledyne API T802 Paramagnetic O
DESCRIPTION
.)
CONC: Samples O
concentration at one minute intervals and stores an average
2
every hour with a time and date stamp.
calibration HOLD OFF are not included in the data. By default, the last 800
hourly averages are stored.
PNUMTC: Collects sample flow and sample pressure data at five-minute
intervals and stores an average once a day with a time and date stamp. This data
is useful for monitoring the condition of the pump and critical flow orifice
(sample flow) and the sample filter (clogging indicated by a drop in sample
pressure) over time to predict when maintenance will be required. The last 360
daily averages (about 1 year) are stored.
CALDAT: Data channel logs new slope and offset of O
time a zero or span calibration is performed and the result changes the value of
the slope (triggering event: SLPCHG). The O
calibration value was stable) as well as the converter efficiency (for trend
reference) are also stored.
This data channel will store data from the last 200 calibrations and can be
used to document analyzer calibration; it also is useful for detecting trends in
slope and offset (instrument response) when performing predictive
diagnostics as part of a regular maintenance schedule.
DEFAULT
SETTING
"NONE"
ATIMER
(PMTDET)
000:01:00
(1 hour)
100
OFF
OFF
3
Readings during calibration and
Analyzer Operation Manual
2
SETTING RANGE
Up to 6 letters or digits
Any available event
(see Appendix A-5).
Any available parameter
1
(see Appendix A-5).
000:00:01 to
366:23:59
(Days:Hours:Minutes)
1 to 30,000 (max), limited
by available storage
space.
OFF or ON
OFF or ON
ON
OFF or ON
measurements each
2
stability (to evaluate if the
2
07275B DCN6418
1
2