Teledyne T802 Operation Manual page 248

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

A Primer on Electro-Static Discharge
13.4.2.5. PACKING COMPONENTS FOR RETURN TO TELEDYNE API TECHNICAL
SUPPORT
DO NOT
NEVER
component/assembly directly. This includes
pack, Styrofoam peanuts, open cell foam, closed cell foam, and adhesive tape
 DO NOT use standard adhesive tape as a sealer. Use ONLY anti-ESD tape
Note
If you do not already have an adequate supply of anti-ESD bags or containers
available, Teledyne API's Technical Support department will supply them
(see Section 11.9 for contact information). Follow the instructions listed
above for working at the instrument rack and workstation.
246
Teledyne API T802 Paramagnetic O
1. Opening the outer shipping box away from the anti-ESD work area
2. Carry the still sealed ant-ESD bag, tube or bin to the anti-ESD work area
3. Follow steps 6 and 7 of Opening Shipments from Teledyne API's Technical
Support above when opening the anti-ESD container at the work station
4. Reserve the anti-ESD container or bag to use when packing electronic
components or assemblies to be returned to Teledyne API
Always pack electronic components and assemblies to be sent to Teledyne API
Technical Support in anti-ESD bins, tubes or bags.
CAUTION – ESD HAZARD
use pink-poly bags.
allow any standard plastic packaging materials to touch the electronic
Never carry the component or assembly without placing it in an anti-ESD bag or
bin.
1.
Before using the bag or container allow any surface charges on it to
dissipate:
 If you are at the instrument rack, hold the bag in one hand while your wrist
strap is connected to a ground point.
 If you are at an anti-ESD workbench, lay the container down on the
conductive work surface.
 In either case wait several seconds.
2. Place the item in the container.
3. Seal the container. If using a bag, fold the end over and fastening it with anti-
ESD tape.
 Folding the open end over isolates the component(s) inside from the
effects of static fields.
 Leaving the bag open or simply stapling it shut without folding it closed
prevents the bag from forming a complete protective envelope around the
device.
Analyzer Operation Manual
2
,
but is not limited to, plastic bubble-
07275B DCN6418

Advertisement

Table of Contents
loading

Table of Contents