Section II Operating Instructions

3.3. CONNECTIONS AND SETUP

3.3.1. ELECTRICAL CONNECTIONS

Note
3.3.1.1. CONNECTING POWER
High Voltages are present inside the analyzers case.
Turn OFF analyzer power before disconnecting or connecting PCAs, wiring
harnesses or electrical subassemblies.
Power connection must have functioning ground connection.
Do not defeat the ground wire on power plug.
Do not operate with cover off.
The T802 analyzer can be configured for both 100-130 V and 210-240 V at either
47 or 63 Hz.
To avoid damage to your analyzer, make sure that the AC power voltage
matches the voltage indicated on the Analyzer's serial number label tag (See
Figure 3-4) before plugging the T802 into line power.
32
Teledyne API T802 Paramagnetic O
This section presents the electrical (Section 3.3.1) and pneumatic (Section 3.3.2)
connections for setup and preparing for instrument operation.
To maintain compliance with EMC standards, it is required that the
cable length be no greater than 3 meters for all I/O connections,
which include Analog In, Analog Out, Status Out, Control In,
Ethernet/LAN, USB, RS-232, and RS-485.
This section presents the electrical connections for AC power and
communications.
W
ARNING
Attach the power cord to the analyzer and plug it into a power outlet capable of
carrying at least 10 A current at your AC voltage and that it is equipped with a
functioning earth ground.