Manually Configuring The Network Ip Addresses - Teledyne T802 Operation Manual

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

Teledyne API

6.3.2. MANUALLY CONFIGURING THE NETWORK IP ADDRESSES

SAMPLE
RANGE=100.00 %
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
PRIMARY SETUP MENU
SETUP X.X
CFG DAS RNGE PASS CLK
SECONDARY SETUP MENU
SETUP X.X
COMM
VARS DIAG
07275B DCN6418
There are several circumstances when you may need to configure the interface
settings of the analyzer's Ethernet card manually:
your LAN is not running a DHCP software package
the DHCP software is unable to initialize the analyzer's interface
you wish to create a static IP (recommended)
1. Connect a cable from the analyzer's Ethernet port to a Local Area Network
(LAN) or Internet port.
2. Access Ethernet configuration through: SETUP>MORE>COMM>INET.
3. Follow the setup sequence as shown in the illustrations that follow, and edit
the Instrument and Gateway IP addresses and Subnet Mask to the desired
settings. (From the computer, enter the same information through an
application such as HyperTerminal).
O2=XXX.XX
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
(continues in next illustration)
T802 Paramagnetic O
2
SETUP X.X
COMMUNICATIONS MENU
ID
INET
COM1 COM2
SETUP X.X
DHCP: ON
<SET SET>
EDIT
SETUP X.X
DHCP: ON
ON
SETUP X.X
DHCP: OFF
OFF
Analyzer Operation Manual
EXIT
EXIT
ENTR EXIT
ENTR EXIT
ENTR accept
new settings
EXIT ignores
new settings
121