Teledyne T802 Operation Manual page 264

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

Appendix A
Setup Variable
SAMP_FLOW_SET
SAMP_FLOW_SLOPE
SAMP_PRESS_SET
BOX_SET
RS232_MODE
BAUD_RATE
A-12
Error! Unknown document property name.Error! Unknown document property name.
Unknown document property name.Error! Unknown document property name.
Models T802, 802E Appendix A Menu Trees (Reference: 06530C DCN6418)
Numeric
Default
Units
Value
cc/m
120
Warnings:
80–180
1
"Hg
29.92
Warnings:
15–35
ºC
30
Warnings:
8–50
BitFlag
0
115200
Value Range
0–6000
Sample flow set point for flow
calculation and warning limits.
0.5–1.5
Sample flow slope correction
factor (adjusted flow = measured
flow x slope).
Sample pressure set point for
0–100
pressure compensation and
warning limits.
5–60
Box temperature warning limits.
Set point is not used.
0–65535
RS-232 COM1 mode flags. Add
values to combine flags.
1 = quiet mode
2 = computer mode
4 = enable security
8 = enable hardware
handshaking
16 = enable Hessen protocol
32 = enable multi-drop
64 = enable modem
128 = ignore RS-232 line errors
256 = disable XON / XOFF
support
512 = disable hardware FIFOs
1024 = enable RS-485 mode
2048 = even parity, 7 data bits, 1
stop bit
4096 = enable command prompt
8192 = even parity, 8 data bits, 1
stop bit
16384 = enable dedicated
MODBUS ASCII protocol
32678 = enable dedicated
MODBUS RTU or TCP protocol
300,
RS-232 COM1 baud rate.
Enclose value in double quotes
1200,
(") when setting from the RS-232
2400,
interface.
4800,
9600,
19200,
38400,
57600,
115200
Description
8
Error!
07275B DCN6418