Table 5-2: Variable Names (Vars) - Teledyne T802 Operation Manual

Paramagnetic oxygen analyzer
Table of Contents

Advertisement

Section II Operating Instructions

Table 5-2: Variable Names (VARS)

NO.
VARIABLE
0
DAS_HOLD_OFF
1
1
STABIL_GAS
2
TPC_ENABLE
DYN_ZERO
3
Special configuration;
consult factory
DYN_SPAN
4
Special configuration;
consult factory
5
CONC_PRECISION
6
CLOCK_ADJ
7
SERVICE_CLEAR
TIME_SINCE_SVC
8
SVC_INTERVAL
9
1
This VAR only appears if the optional CO
86
Teledyne API T802 Paramagnetic O
DESCRIPTION
Changes the internal Data Acquisition System (DAS) HOLD
OFF timer:
No data are stored in the DAS channels during situations
when the software considers the data to be questionable
such as during warm up or just after the instrument returns
from one of its calibration modes to SAMPLE Mode.
Selects which gas measurement is displayed when the
STABIL test function is selected.
NOTE: It is strongly recommended that this variable
NOT be changed.
ON enables, OFF disables temperature and pressure
compensation
[Automatically adjusts offset and slope of the O
when performing a zero point calibration during an AutoCal.]
[Automatically adjusts offset and slope of the O
when performing a zero point calibration during an AutoCal.]
Allows the user to set the number of significant digits to the
right of the decimal point display of concentration and
stability values.
Adjusts the speed of the analyzer's clock. Choose the +
sign if the clock is too slow, choose the - sign if the clock is
too fast.
Pressing the OFF button to display SERVICE_CLEAR:ON,
followed by pressing ENTR resets the service interval timer
and returns this Var back to its default setting, ready for the
next reset.
Displays time in hours since last service (restarted by the
SERVICE_CLEAR Variable).
Sets the interval in hours between service reminders.
sensor is installed.
2
Analyzer Operation Manual
2
ALLOWED
SETTINGS
May be set for
intervals
between
0.5 – 20 min
O
; CO
2
ON, OFF
response
2
[ON, OFF]
response
2
[ON, OFF]
AUTO, 1, 2,
3, 4
-60 to +60
s/day
ON, OFF
0-50,000
0-100,000
07275B DCN6418
VARS
Default
settings
15 min.
O2
2
ON
[OFF]
[OFF]
AUTO
0 sec
OFF
0
0