Physical Range Versus Analog Output Reporting Ranges - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual
On analyzers with the SO
Output
Channel
A1

A2

As the instrument switches from H
output associated with the gas currently being measured will be active. The reporting range and analog
output for the gas no being measured will continue to report the last valid reading.

4.4.4.2. Physical Range versus Analog Output Reporting Ranges

AUTO mode, see section 4.4.4.6) or operate independently (IND mode, see Section
4.4.4.5). The user may also select between a variety of reporting range spans as well:
EXAMPLE:
A1 OUTPUT: Output Signal = 0-5 VDC representing 0-1000 ppm concentration values
A2 OUTPUT: Output Signal = 0 – 10 VDC representing 0-500 ppm concentration
values.
-H
S multigas measurement option installed the outputs of A1
2
2
and A2 correspond to:
SO
2
Mode
SO
2
SO
2
S mode to SO
2
The output, labeled A4 is special. It can be set by the user (see Section 4.6.9) to output
many of the parameters accessible through the <TST TST> buttons of the units Sample
Display.
Output A3 is not available on the Model T101 Analyzer.
The T101 UV Fluorescence H
H
S concentrations between 0 and 20,000 ppb. The low range is 0 to 2,000 ppb, while the
2
high range is 0 to 20,000 ppb. The proper physical range is determined by the software to
include the maximum measurement concentration selected by the user. Once properly
calibrated, the analyzer's front panel will accurately report concentrations along the entire
span of its 0 and 20,000 ppb physical range.
Because, most applications use only a small part of the analyzer's two physical ranges,
the width of the Model T101's physical range can create data resolution problems for
most analog recording devices.
concentration of SO
is typically less than 500 ppb, the full scale of expected values is
x
only 0.25% of the instrument's 20,000 ppb physical range. Unmodified, the
corresponding output signal would also be recorded across only 0.25% of the range of the
recording device.
The T101 solves this problem by allowing the user to select a scaled reporting range for
the analog outputs that only includes that portion of the physical range relevant to the
specific application. Only the reporting range of the analog outputs is scaled, the physical
range of the analyzer and the readings displayed on the front panel remain unaltered.
NOTE
SO
– H
S
2
2
Mode
SO

2
H
S

2
mode and back, only the reporting range and analog
2
S Analyzer has two hardware physical ranges that cover
2
For example, in an application where the expected
Operating Instructions
H
S
2
Mode
H
S

2
H
S

2
67